Ion Pairing reagent/Phloxine B
Posted: Wed Mar 20, 2019 6:01 pm
by Andres760
I'm testing for Phloxine B using 55% ACN with 0.03M TBAB at 548 nm. I'm getting peak splitting, could this be caused by too much ion pairing reagent, and what are the effects of using too much?
Re: Ion Pairing reagent/Phloxine B
Posted: Wed Mar 20, 2019 8:34 pm
by Vlad Orlovsky
How much you are injecting and what is the retention time? You might be seeing effect of large volume injection. Also, mixed-mode columns will allow you to avoid the use of IP reagent (it is attached to the surface of silica)
Posted: Wed Mar 20, 2019 9:49 pm
by Andres760
I'm injecting 20 microliters. I noticed that if I use 65% with 0.01M TBAB, the peaks look completely different, and I cannot tell if it's the ACN or the TBAB.