lower detection limit
Posted: Fri Oct 31, 2008 7:55 pm
by gokul
Please help
an interview Q I couldn't answer can you please provide answer...
What parrameters do we need to change to lower the detection limit by a factor of 10 in a method??
Posted: Fri Oct 31, 2008 8:32 pm
by tom jupille
The general answer is that detectability is ultimately limited by signal/noise ratio. There are only two ways to improve that:
- more signal
- less noise.
Under "more signal":
- inject a 10 x larger volume (this works only if they define detectability in terms of concentration)
- get a 10 x smaller column volume
- get a 10 x longer flow cell
- decrease retention (will help, but not 10 x)
- use smaller particles (will help, but not 10 x)
- find a better wavelength (maybe)
- make a more detectable derivative (maybe)
- find a more sensitive detection technique
Under "less noise":
- improve instrument performance (lamp age, degassing, temperature control) if system not meeting specifications.
- detector time constant (works for short-term noise only)
All of the above may have ramifications in terms of affecting other characteristics of the method, and may not be permissable in the case of a validated method.
thanks
Posted: Sat Nov 01, 2008 10:00 am
by gokul
Thanks Tom...
I thought of those when i'm home but donno exactly... I appreciate your reply
Regards
Gokul
Posted: Tue Nov 04, 2008 4:59 pm
by Anthony Crawshaw
One thing that may also be worth trying is playing with your sampling rate on your detector. This can have quite drastic effects on your s/n ratio.