toluene diisocianate separation problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

5 posts Page 1 of 1
dear all
I am new in LC field.
my task is to monitor 2,4-TDI and 2,6-TDI in air sample. I use: eluent: 0.2M ammonium-acetate buffer, pH=6-6.2 with acetic acid; aceto-nitrile 70:30.
Column: Philips cyano 250 x 4.6, 5um
Flow: 1ml/min
Det: fluorescence, ex: 240nm, em: 370nm
This column has not been used for a long time and while measuring the baseline fluctuates. How can I stabilize the baseline? Furthermore, I have different concentrations of 2,4 , 2,6-TDI and on the chromatogram there are 6 peaks and I do not know which belongs to them. Unfortunately I don't have pure standards.
What would you suggest to identify them? Thank you.

You can but 2, 3 and 2, 6 TDI Stds from Aldrich
Good judgment comes from bad experience, and a lot of that comes from bad judgment.

When you say that the baseline fluctuates, exactly what do you mean? Is the fluctuation regular (ie. every 30 sec) or is it constantly increasing or is it a more random fluctuation?

Hi sassman
It is a random fluctuation. I don't know if it is a mistake of the column or anything else. After a peak the baseline drops down below zero and stays there until next peak. Maybe there is a bubble in detector because I'm only using ultrasonic bath to eliminate the bubbles from the eluent.
Sorry for my english. :oops:

You need to check if your LC system works properly. I would bypass the column and run the same method with injection. If the similar fluctuation is observed, you will need to trouble-shoot the instrument.
Did you use premixed mobile phase? If not, try to so. Malfunctioned proportining valve affects detection.
To verify the column, it is better to run the same method on a known good instrument.
As for the application, I am attaching a link which leads to the same application we did on a Acclaim Mixed-Mode HILIC-1 column: http://www.dionex.com/en-us/images/pdf- ... hod-42.pdf
Xiaodong Liu
5 posts Page 1 of 1

Who is online

In total there are 9 users online :: 2 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: John Guajardo, Semrush [Bot] and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry