JA » Fri Sep 19, 2008 11:08 pm
reload..reload..reload..reload..reload..reload..reload..reload..
reload..reload..reload..reload..reload..reload..reload..reload..
I'll put my hand in and have a go at this one while we wait for the resident expert on Waters packing materials and prep LC
I haven't come across a standardised test for column capacity and thus I don't think that the information you have requested is actually out there. Sure, one can acquire data on the physical properties of various stationary phase packings but it will not include reference to their sample capacity. Users often assume that this capacity increases in proportion to variables such as the particle surface area, ligand chain length or carbon load which is not necessarily the case when comparing different packings,
e.g. a C18 from two manufacturers.
You can calculate the sample capacity of a given column assuming you have already determined the capacity of one which encorporates the same column chemistry. This is one of our fundamental approaches in preparative chromatography: determine the loading on an analytical column and scale up in proportion to the change in column volume. If we decrease the particle size, I'd expect the capacity should go up in proportion to the specific surface area assuming that the ligand bonding density remains unchanged.
One way to measure the column capacity would be to spike a known concentration of a test probe into the mobile phase and measure the breakthrough time. I'm not sure if this will yield publication-quality data but it should let you make informed comparisons.
In summary, determining the sample capacity (loadability) is very much in your own hands.
It is entirely application specific. Preparative chromatography is not governed by the appearance of pretty narrow and symmetrical peaks and is usually performed in overload conditions. The only criteria for a successful project is one where the product is recovered with satisfactory yield and purity.