Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
i recently calibrated an HPLC instrument. in order to pass the temperature precision test i needed a capacity factor between 0.9 and 1.8. the run i carried out were all the same except the temperature of the column was altered. how does this experiment prove that the temperature is precise? is that even what it's trying to prove?
MestizoJoe
Analytical Chemist and Adventurer
Venture Industries
Spider-Skull Island

I find it interesting that you can show the precision of a column heater by monitoring the variation in retention time for some chemical probe.

I wonder what upper and lower limits of temperature give you k' values from 0.9 to 1.8 in your method, all else being equal. If I recall correctly a rule of thumb is approximately 1% change in RT for a 1°C change in temperature, so this looks like quite a wide working range.

I would expect that temperature accuracy and precision would be better demonstrated using a certified thermocouple. With all the other variables in any particular HPLC methodology I don't see how retention time variation can be attributed to temperature alone.

you experiment is interesting.

for it to be anything close to valid, does it show the translation of the K' value to temperature values?

Maybe it is just a misnomer for a test of temp. on retention time?

This looks like the test for temperature accuracy from Waters.
I think waters allows 5°C deviation from the set temperature, so this is also a large range.

One can verify the trueness of the test by setting the temperatures, verify by an external temperature probe and inject. Do this for 30°C, 35°C, 40°C, 45°C and 50°C and you know wether the test is correct or not.
(And please let us know)

All the best

Ace

This is simply a van't Hoff plot (lnk vs 1/T). In most cases pure reversed phase separations will give a linear relationship, so one can indeed calculate temperature from retention factors.
This relationship can become very significant if the samples undergoes a structural change at a certain temperature. Denaturing HPLC (dHPLC) is such a case where separation temperature control down to 0.1 or 0.2°C becomes necessary.
--
Robert Haefele
6 posts Page 1 of 1

Who is online

In total there are 27 users online :: 1 registered, 0 hidden and 26 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot] and 26 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry