Fluctuating pressure

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi all,

I have this problem when running Hilic column: the pressure is about 94 bars, but the reading is rather fluctuating. I illustrate in the picture below.
Have anyone met this problem ?.
Thanks


Image

That amount of pressure fluctuation could be normal for your LC system. Did it used to be smaller? With pressure fluctuations, I always try to clean the check valves first by sonicating in isopropanol. Sometimes a small particle can get in the valve and prevent it from seating properly. Bad seals can also cause pressure issues. But, like I said, this could be normal for your LC. We have a number of LCs that fluctuate up to 25psi or almost 2bar in normal operation.

I don't use that type of column, but at 94 bar, 2-3 bar fluctuations can be normal, 1 looks pretty good.
3 posts Page 1 of 1

Who is online

In total there are 12 users online :: 2 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry