Robustness for gradients

Posted: Mon Oct 20, 2008 5:50 am
by michaelcarolus
Morning All

I'm busy with a method valiation and I need to do robustness on a gradient method. I was thinking of doing the following for robustness:
1. changing the column (using an equivalent)
2. running the gradient on a different HPLC system.
3. change the gradient times by ± 1 minute.

Are there any other parameters that I could change? Any suggestions would be appreciated.

Thanks
Michael Carolus

Posted: Mon Oct 20, 2008 12:53 pm
by lmh
The thing that is likely to vary most is solvent preparation. If you know other labs are likely to use different columns, by all means change the column and see what happens, but I personally would feel this sort of change goes beyond the scope of robustness. On the other hand, if a slight pipetting error with a gradient additive, or a misunderstanding about how to make 1% acetic acid in water, causes a 50% change in retention times, then there might be trouble ahead.

Re: Robustness for gradients

Posted: Mon Oct 20, 2008 6:28 pm
by Noser222
2. running the gradient on a different HPLC system.
Dwell volume and type of pump systems (binary pump vs. proportioning valve) are the issues there. If you account for those then the experiment where you change gradient time by 1 minute sort of becomes irrelevant.

Posted: Mon Oct 20, 2008 9:30 pm
by danko
Testing the â€

Posted: Tue Oct 28, 2008 7:30 am
by michaelcarolus
Thanks for the comments/suggestions. I will take them into account when I do my robustness.

thanks again
Michael