Shifting Retention times/Peak areas

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi,

I have an Agilent 1100 system, I am running a gradient mobile phase through a c18 column. I am failing system suitability repeatedly. When I examined my start/End pressures I am experiencing significant drops in pressure - after column conditioning the pressure seems stable at 160 bar, but after a few injections the pressure drops to as low as 115 bar. I have also witnessed the pressure climbing back up at the end of the run for the remaining Supelco injections - any ideas?

It is normal for the pressure to drop during a gradient run since the organic part of your mobile phase has lower viscosity than water. Do you have a sufficient equilibration period at the end/beginning of your run? If you can give more details about your method, it might help us to diagnose the problem.

My gradient is as follows:

Solvent A Water
Solvent B Scetonitrile
Sovent C 1% TFA

Time (Minutes)----Flow (mL/Min.)---------%A----%B----%C
N/A------------------------1.25----------------65----25-----10
1.0------------------------1.25----------------65-----25-----10
25.0-----------------------1.25----------------45-----45-----10
30.0-----------------------1.25----------------45-----45-----10
35.0-----------------------1.25----------------65-----25-----10
40.0-----------------------1.25-----------------65----25-----10
120.0----------------------0.00-----------------65----25-----10

Conditions are:

1.25mL/min
30min run time
10min next inject delay

Pressure changes we have witnessed are as follows:

---------------------------------------Pressure (Bar)
---------------------------Start Pressure End Pressure
Equilibration - 40mins---------161--------------162
Column Condition 160mins---161--------------163
Standard Inj 1------------------161-------------144
Standard Inj 2------------------160-------------141
Standard Inj 3------------------157-------------139
sample matrix inj 1------------155-------------137
sample matrix inj 2------------152--------------135
sample matrix inj 3------------150-------------135
sample elution inj 1-------------150------------134
sample elution inj 2------------149------------134
sample elution inj 3------------149-------------134
sample @ 5micrograms--------147-------------133
sample @ 5micrograms--------146-------------113
sample @ 5micrograms---------122-----------113
sample @ 5micrograms----------119-----------112
sample @ 5micrograms----------118-----------109
sample @ 5micrograms----------115-----------131
Sample @ 3micrograms-----------137----------121
Sample @ 3micrograms-----------130----------118
Sample @ 3micrograms----------146-----------128
Sample @ 7micrograms-----------133----------116
Sample @ 7micrograms---------121------------131
Sample @ 7micrograms------------137----------117
Standard Inj 1------------------123--------------136
Standard Inj 2-------------------151--------------139
Standard Inj 3------------------155-------------140
3 posts Page 1 of 1

Who is online

In total there are 31 users online :: 1 registered, 0 hidden and 30 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 30 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry