Here are my conditions:
Varian 3800
1 ul injection
200 fold split
250 degrees injector
270 degrees detector
Range 12
Attenuation 1
2.0 ml/min flow rate, with constant flow mode enabled
Wax column 30 m x 0.25 mm x 0.25 um film
170 degrees C hold for 2 minutes, to 240 degrees at 2.9 degrees/min. Hold for 4 minutes.
2 mm internal diameter insert, no glass wool.
injection speed 5 seems to work, but not consistently.
According to the restek backflash calculator (http://www.restek.com/info_calcs.asp)
at 1.0 ul injection with isooctane and a 54 mm insert, I'm at the upper limit of the capacity of the insert.
I then lowered the injection volume to 0.5 ul, but that was too low. However, when I did a 3 rep at 0.8 ul, the area counts were all over. For one peak, of the same standard under the same conditions, I got:
14743
25122.9
57813.3
I currently have the injection mode set at User, and just reset the septum purge flow to 5 ml/min.
I thought I had the issue resolved by shooting at 0.8 ul, within the capacity of the liner, but that did not work.
Any and all suggestions are quite welcome at this point!