What causes peptides to have longer retention on a C4 vs C18

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

11 posts Page 1 of 1
I have a peptide I'm trying to isolate. It was more highly retained on the Vydac C4 column than it was on a Zorbax C18 column (i.e., required higher % acetonitrile to elute).

Firstly, I'd like to ask if this is "real" (i.e., does this mean there is likely something wrong with the C18 column, and if it were new, would it be less likely to occur?)

How unusual is this?

What, physically, would cause this?

not unusual, could just be pore size...

What if the pore size were the same?

Depending on pH of your mobile phase you can see more ion-exchange interaction on C4 in addition to reverse phase, as on C4 phase silanols are accessible than on C18.

Pipettemonkey, are the dimensions of the Vydac and the Zorbax columns identical?
I can imagine that the Zorbax is a 2.1 mm (inner diameter) and maybe shorter as well.
Whereas the Vydac is the more traditional 4.6 mm. And if you applied the same flow rate on both columns, then you have the explanation.

Best Regards
Learn Innovate and Share

Dancho Dikov

Pipettemonkey, are the dimensions of the Vydac and the Zorbax columns identical?
I can imagine that the Zorbax is a 2.1 mm (inner diameter) and maybe shorter as well.
Whereas the Vydac is the more traditional 4.6 mm. And if you applied the same flow rate on both columns, then you have the explanation.

Best Regards
Both columns are the same ID. I use the same flow rate (1 mL/min) and an acetonitrile gradient with 0.1 % TFA.

Same length as well?
Is the Zorbax column also a Poroshell variant?

Best Regards
Learn Innovate and Share

Dancho Dikov

Same length as well?
Is the Zorbax column also a Poroshell variant?

Best Regards
They are different lengths. The C4 is 15 cm, the C18 is 10 cm

There you go :)
Learn Innovate and Share

Dancho Dikov

There you go :)

I don't think so. The difference in retention is too great to be accounted for by column volume.

Have you seen the same patten with any peptide or just the one that you are trying to isolate. Maybe there is something to do with the sequence/structure of the specific peptide that you are trying to isolate. In any case if you are not sure about the status of your Zorbax column (i.e. it is not a new one) maybe your results are not representative of what is really happening...

Having said that I know that the Zorbax 300A material has a pretty small specific surface area (i.e. 45 m2/g) while the Vydac I think it is higher (although I do not remember the exact number. The phenomenex Jupiter in comparison for 300A has 175 m2/g.
11 posts Page 1 of 1

Who is online

In total there are 22 users online :: 0 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 22 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry