I was wondering if anyone can provide advice in regards to the difference in chromatography that I am seeing between analysis performed on a Perkin Elmer Autosystem XL GC with Headspace Sampler HS40 and our newer Agilent 6890N with G1888 Headspace Sampler.
I am running a method for determination of ethanol in drug product. This method has been in use for a number of years and has been run on the Perkin Elmer with out problems.
The conditions currently employed for analysis on both systems are as follows:
N2 gas at 8 psi
Loop temp 90C
Equilibration 30 min
Pressurise time 2.0 min
Injection time 0.1 min
Air flow 450 mL
Hydrogen flow 45 mL
Inlet 90C
FID temp 260C
Column temp 40C
Hold 10 min
Ramp rate 15C/min
Final 220C
Hold 2 min
The main problem we are encountering is in regards to the recovery we are seeing on the Agilent system. Our recoveries on the Agilent are consistently around 130%. Where as our analysis performed on the Perkin Elmer will consistently give us recoveries of around 110%. All of these are consistent regardless of sample size (weight). What could account for this difference in recoveries? We have run numerous accuracy samples and always have the same result. Does anyone have any advice as to what could be causing this difference in recoveries? We are in a position where altering parameters and conditions is an option and are open to all suggestions.
If any further information is required for you to help answer the question I am happy to provide more information.
On a side note I have noticed that the Agilent system results in poor standard precision (approximately 3.0% where as the Perkin Elmer is approximately 1.0% at worst) has anyone else working with an Agilent system encountered this?