No peaks (PCBs) for change of column with ECD detector
Posted: Fri Oct 10, 2008 6:43 am
Hello my name is Juan Manuel, I´m working with PCBs in soils and sewage sludge.
I´ve worked with a column VF-5ms 30 meters and 0.32 diameter. The method is:
- Gas Make-up and carrier N2
- Pressure 5.5 psi
- Mode injection splitless 1.5 minutes after split 1:40
- Make up 40 ml/min
- Temperature injection 250ºC, detector 300ºC and oven 50ºC for 3 minutes, 20ºC/min up to 180ºC, 2.5ºC/min up to 280ºC after 10.5 minutes.
My chromatogram has bad resolution for pcb 28 and 52 I want to improve it.
And now I have a column CP-SIL 8CB PCB FS 50 meter and 0.25 diameter. I am trying translate the other method with this column but at this moment I can´t see nothing Peak.
My new conditions are:
- Pressure 24 psi the other similar. No peaks.
I´ve realise injection without split, more pressure 40 psi, more sample, and I can´t see the peaks.
¿What´s problem I can have?
- Column bad situation?
- The make-up isn´t good for this conditions?
I did the calculations whit software of translate.
Thank you very much for you attentions
- Pressure 5.5 psi
- Make up 40 ml/min