Without having performed any of the work myself, and without seeing the actual chromatograms, the information is currently limited thus:
Using a water/acetonitrile isocratic method, a reasonable purity (>95%) is indicated on a Zorbax StableBond C18 column. The same mobile phase, when used on an Ascentis Express C18, displays additional early-eluting peaks which reduce the apparent purity by approximately 10%. Gradient runs on the Zorbax give results consistent with the isocratic data, while such a 'neutral' gradient has not been utilised on the Ascentis Express.
On the Ascentis Express column, the usage of acidic (H2O/MeCN/TFA) or basic (amm bicarb/MeCN) mobile phases appear to promote degradation and lead to a lower apparent purity than observed in any of the preceding work.
Said colleague reasoned that the Ascentis Express possesses more acidic (Si-OH) and basic (Si-O-) sites in comparison to the Zorbax StableBond support, particularly since production of the latter employs the use of bulky endcapping reagents to prevent C18 ligand hydrolysis. A second colleague then assigned Ascentis Express to be the more acidic stationary phase.
Is there an expectation that an acid labile compound would display a lower apparent purity on a stationary phase possessing a larger degree of surface silanols? Are silanols acidic in the sense they can donate a proton to a particular analyte being chromatographed? Do we even know which of the two columns would have more surface silanols and do we know which is likely to have more higher energy sites?
I appreciate that the information provided doesn't present a complete comparison of the two columns (I might be able to dig more up if it will assist) but I'd like to hear your thoughts.