already removed the column and there is still no flow.
any help appreciated.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
That means you have fluid flowing down to the pump. That is good.thanks everyone. I am able to narrow it down now. I Turned off the system and when I unscrewed the thin plastic tube, from multichanel mixing valve into the pump, I saw mobile phase flow coming through.
The same is true for the inlet valve, and the inlet valve fails more frequently than the outlet valve. There is a simple test for a leaky inlet valve. When the pump is delivering mobile phase to the system from one channel and the pressure is intensively pulsating, as described above, allow an elongated air bubble (ca. 5-10 mm in length; not just tiny spherical bubble) to enter the inlet line. Then see how the bubble is moving along the tube. If the forward movements of the bubble are alternating with periodic stops (in the rhythm of the pump), the inlet valve is OK. If the forward movements of the bubble are alternating with more or less slow backward movements, the inlet valve is not closing properly.If pressure build but falls again with the rhythm of the pump, then the outlet check valve is not closing
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.