Page 1 of 1

Derivatization

Posted: Fri Feb 01, 2019 10:26 pm
by dlbenach
Hello,

I'm trying to derivatize a standard solution of acetic acid and ethylene glycol in acetonitrile (ACN). I'm taking 100 uL of the solution and 5 uL of BSTFA w 1%TCMS into a 20 mL headspace vial and heating at 80 C for 15 min before injecting. I ran a blank, 100 uL of ACN and 5 uL BSTFA w TCMS, and saw 3 large peaks in addition to the ACN peak. Here is the chromatogram.

Image

Would you expect to see the 3 large peaks from the BTSFA w TCMS? I do not have any reference method or chromatogram to compare to and I haven't done this before. Just to make sure the peaks are not coming from the ACN, I injected just 100 uL of ACN. Here is that chromatogram. The 3 peaks are not preset.

Image

Thanks for your replies,

Dave

Re: Derivatization

Posted: Sat Feb 02, 2019 11:16 am
by Aldijkhuizen
Maybe mono and di tns of glycol? Try derivatising seperate. Do you have acces to gcms? Then you can identify

Re: Derivatization

Posted: Mon Feb 04, 2019 1:17 pm
by Peter Apps
Entirely possible.

Peter

Re: Derivatization

Posted: Mon Feb 04, 2019 2:24 pm
by dlbenach
The first chromatogram posted is of only 100 uL of ACN and 5 uL of BSTFA w 1% TCMS in a headspace vial. The vial did not contain any glycol or acetic acid. I am wondering if you normally see large peaks from the derivatizing agent?

Thanks,

Dave

Re: Derivatization

Posted: Mon Feb 04, 2019 3:30 pm
by Consumer Products Guy
Maybe mono and di tns of glycol? Try derivatising seperate. Do you have access to gcms? Then you can identify
We used BSTFA with TMCS (trimethylchlorosilane, not TCMS) a ton. But we always used DMF or pyridine as the solvents (and never did headspace). We ALWAYS got complete derivatization of ethylene glycol, propylene glycol, glycerin, etc.