ASTM D-3606 Column Set
Posted: Tue Sep 30, 2008 7:11 pm
by TheLabDirector
Greetings, can anyone tell me what column sets you are using for D3606. I've tried 6 sets of RESTEK's D3606 sets and they are totally inconsistent. Supelco is having trouble with there OV-101. I am definately SOL right now. Any help would be greatly appreciated.
Posted: Wed Oct 01, 2008 1:25 pm
by AICMM
LabDirector,
You will note that Restek's application note shows a very small injection volume and an FID which means column overloading is not nearly as big an issue. Having said that ,I have installed a D3606 set from them and it works great but I am only shooting 1/2 uL. I will dig around and see if I can find any further info. This would be one for Rod (O' where art though.)
Best regards.
Posted: Thu Oct 02, 2008 2:30 pm
by gov10
Lab Director,
I have used the Restek D 3606 column set and I can't say enough about the performance. Relative to column overload, I typically inject 1.0- 1.5ul with no problem. The ethanol/benzene resolution is typically > 3.5.
Regarding the issues you had with the Restek columns, I suggest you give them a call, ask for Barry, I can't remember his last name.