what is cut off for tetraalkylammonium sal (HPLC) ?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
I m sorry friends, but I have one doubt.....I need to analysis for organic acids, but the literature recomends TBA....I know the advantages when you use this ion par, but I have been problems because the low wavelength (2100 nm UV detector) for 0,002M TBAI pH 3,5 mobile phase??
:(

You can purchase a version of this reagent that is good for use down to 200 nm from Waters: low-UV PIC A, part number WAT084189
Now, I need to check the specification about the PIC-A Waters, but are you now the criteria (TBA/ cloride/ phosfate or sulfate) for ionic compounds....
Dr. Jose Carlos Penteado

Just to clarify, if I recall correctly, Waters PIC A ( tetrabutylammonium ) comes as two forms.

"Low UV PIC A" ( tetrabutylammonium hydrogen sulphate ) which can be used down to 200nm, and "PIC A" ( tetrabutylammonium phosphate ) which can be used down to 240 nm. Those cutoffs apply to the recommended concentrations.

Make sure to look at the correct product specifications. There are pure forms of tetrabutylammonium salts available from Romil, Sigma-Aldrich, and other suppliers.

Bruce Hamilton

Ion-pair reagents have been sold by Waters since the 1970s, and the high-purity form has been sold since at least the early 1980s. Waters was a leading supplier of these reagents, whose use goes back to the work of Brian Bindlingmeyer, who was then a Waters employee, and his team.
This question about TBA...started with the article Determination of glycolic acid in cosmetic products by solid-phase extraction and reversed-phase ion-pair high-performance liquid chromatography...this work have been used TBAI but 210 nm UV detection.....but I know this salt is no good, when used low UV....sometimes....is possible
Dr. Jose Carlos Penteado
6 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], John Guajardo and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry