how should i start my method development for HPLC
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
4 posts
Page 1 of 1
if having no information(only solubility/moleculer structure) on analyte than how should i start my HPLC method development
-
- Posts: 2916
- Joined: Mon Aug 30, 2004 10:19 pm
In principle, run a gradient from 5% acetonitrile to 90% acetonitrile. If you suspect that you have ionizable compounds in your sample, add an acid to the mobile phases: about 5 mM phosphoric acid (check the miscibility/solubility in the 90% acetonitrile/10% 5 mM phosphoric acid) for UV detection or 5 mM formic acid for MS detection. With UV, monitor at a low wavelength (~215 nm) to see everything that elutes from the column.
See where the stuff is coming out and narrow the gradient or go isocratic.
See where the stuff is coming out and narrow the gradient or go isocratic.
-
- Posts: 239
- Joined: Tue Nov 09, 2004 3:06 pm
first look at the detection possibilities -UV is easiest, if it has a chromophore.
then at solubility- if its water/methanol/ACN soluble try Uwe's way,
if its only soluble in apolar compounds, go for normal phase
Alex
then at solubility- if its water/methanol/ACN soluble try Uwe's way,
if its only soluble in apolar compounds, go for normal phase
Alex
-
- Posts: 156
- Joined: Tue Sep 07, 2004 8:23 pm
I am aware of a lot of labs where the start to inject the compound on a HILIC column.
If that doesn't work, then they try RP columns.
The reason for the strategy is that HILIC typically will give better sensitivity in MS or ELSD, faster and better sample preparation, and retention also for impurities. So the gain is considerable.
If that doesn't work, then they try RP columns.
The reason for the strategy is that HILIC typically will give better sensitivity in MS or ELSD, faster and better sample preparation, and retention also for impurities. So the gain is considerable.
------------------------
Merck SeQuant AB
http://www.sequant.com
Merck SeQuant AB
http://www.sequant.com
4 posts
Page 1 of 1
Who is online
In total there are 12 users online :: 0 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 12 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 12 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science