I'm trying to validate a hplc method to determine the sibutramine hcl dissolution sample.
The condition was :
Dissolution medium : acetate buffer pH 4.00
Standard solution : 15 ppm of Sibutramine HCl in dissolution medium.
Column : Phenomenex Luna 5micro C18(2) 250 x 4.6 mm
Mobile phase : Acetonitrile:0.025M Phosphate buffer with 0.1% TEA (40:60)
Column temp : 30 degree Celcius
Flow rate : 1.2 ml/minute
The problem is, the retention time keeps decreasing. I equilibrated the column for about 20 minute before injecting the sample. My first injection had the retention time about 30 minute. the second injection got the 15 minute RT, down to the third injection the RT had become 3.5 minute!
Meanwhile the pressure dropped from 117 bar down to 100 bar.
I already mix my mobile phase and put it into one container to anticipate the pump proportioning problem.
Can someone please give me a clue of what's probably going on here?