JGK » Mon Sep 22, 2008 6:56 pm
We have been thinking a lot about high pressure HPLC in our laboratory.
I am well aware that when you go to very high pressures you have to reduce the diameter of the column to minimize the radial temperature gradients (caused by viscous heating).
Here's my question: How high can you go while still using a standard 4.6 mm column? Can you go to 7000 psi...8000 psi....etc. And how about with a 3.2 mm ID column.
This is important because we are trying to figure out what pressures we can work at without having to take on all the difficulties of extra-column effects.
Thanks
You will be lucky if you can get up to 5000 psi on a conventional HPLC. Backpressure will depend on particle size. Also with some of the older columns and packings you could collapse the stationary phase causing voids and poor chromatography.
I had a method which involved a short 50 mm, 1.7µm column with a mostly (80%) organic MP operating at 0.5 mL/ min. th backpressure was 5200 psi.
I could only get the flow to 0.5 mL/min in 0.1 mL/min stages allowing the flow to equilbrate each step before increasing.
This was done on a Thermo system with a P4000 pump If I had tried it on our Waters system, I dont think it would have handled it.