Hi,

I’m trying to develop a method to separate fusel oil on alcoholic beverage, the main component are:
Acetaldehyde
Methanol
Methyl acetate
n-propanol
2 butanol
Ethyl acetate
Iso-butanol
n-butanol
isoamyl alchol
Active amyl alchol
Sec-butyl acetate
n-amyl alchol

usually this type of analysis is done with a packed column, I would like to do it with a capillary column to don’t change my GC (I’m using it for other method that required capillary!)

I tryed with different column CARBOWAX, the last was Supelcowax 10 60m, 0.32 but no matter what I change I always had problem at the end if the chromatogram ( no resolution at all on the amyl alcohol).

I found online an application that was perfect for me, from Supelco #164, with SPB-20, the only difference is that they run with nitrogen and I use helium, on the application is written that only the analysis time will change but is not really like that, all the chromatogram change and I completely lost resolution between acetaldehyde, methanol end ethanol, the rest of the chromatogram is not bad.

any suggestion as a set up in a method translation like that? ( I try different temp, flow but no much improvement ).
Is there any other column more appropriate to do this type oof analysis?
[/list]

Hello mmegliol; I use to run this method for Rum a couple of year before in a CARBOWAX capillary column of 25 mts; the resolution was good but the problem was in the separation of Acetaldehyde - Methanol. I dont remember very well the chromatographic contidition but the Method start at 20C and 12 psi (hard to reach that temperature, and sometime I deceided to cheange to 24 C and 10 psi more reliable) and then continue with a slow ramp.

what are the chormatographic condition u found in the internet?

it's funny becasue on a carbowax 60m I have no problem to separate acetald and metoh, but on the SPB-20 is a disaster!

on the run with nitrogen they used 40 hold 4 min, ramp4 by min until 110 hold.
velocity at 11 cm/sec.

thanks for your help.
3 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry