Evaluation of self packed MPLC column
Posted: Fri Sep 12, 2008 8:26 pm
Hello
I'm not very experienced in preparative chrom and would like to ask you some questions.
(I think some questions/doubts are arising cause of too much HPLC thinking...
)
We've packed a MPLC column (26x230 + precolum 11ml) with silica gel Si60, 15-40µm (Merck), total packing:
64g silica, slury packing, compacted up to 38bar.
Now we would like to evaluate the quality of the column.
We prepared a test mix with naphthalene/anthracene (5/0.5 mg/ml hexane) and run it with hexane 100% @10 and 20 ml/min
The result was that the resolution and peak shape was "poor" with injection of 1 or 2 ml. Plate count was about N=900 which would lead to a plate height of H=255 µm (corrs. h=6.4 (for 40µm particles))
Injecting anthracene for its own in a 1:10 dilution gave better peakshapes and about 2000 plates (H=115µm, h=2.88)
Beside of this the k' values don't correspond to those obtained by short TLC evaluation (Merck HPTLC plates, Si60).
Don't have the values by hand right now (think the Rf were about 0.5 and 0.7). On MPLC the first peak came out virtually with t0, followed by anthracene.
We then tried another test mix with phenone homologues (aceto-, propio-, butyro-) and hexane/t-butyl methyl ether 95/5.
Propio- and Butyro couldn't be separated good enough, so we prepared a solution of only aceto- and butyrophenone, with some toluene as t0 marker, dissolved in hexan 100%. b(phenones)= about 12mg/ml total.
Injecting of 1 ml gave a satisfying resolution, the k values were about 0.29 and 0.75, t0 was 4.8min, flow 21 ml/min. N was about 1200 (for the second peak.)
Injection of 0.5 ml the plate count rises to about 1800.
Injection of 2 ml, the plate count droped to about 800.
Beside of the plate counts there are these small k values. Evaluated by TLC we would have excpected k's of 1 and 1.6...
Now I wonder if someone can unravel these mysteries...
What effects do I see?
- Volume-overload, even if sample is disolved in eluent? Already with 1 ml or 2 ml?
- Bandspreading before and after column? (bandspreading volume was evaluated to be about 12 ml (w4.4% method))
Which plate count is "true", should be used for the estimation if a separation can be done with this colume?
How do/can I get these significant deviations in k values for silica (TLC vs MPLC)?
Where should I further "shoot troubles", column or flowpath?
Are these observations "normal" for MPLC?
What are some rule of thumbs for MPLC?
- reduced plate heights?
- injection volumes related to the column volume?
- mass load /g silica
- tests for evaluation, if the column can be used or bette should be repacked?
Thank you for your replies
I'm not very experienced in preparative chrom and would like to ask you some questions.
(I think some questions/doubts are arising cause of too much HPLC thinking...
We've packed a MPLC column (26x230 + precolum 11ml) with silica gel Si60, 15-40µm (Merck), total packing:
64g silica, slury packing, compacted up to 38bar.
Now we would like to evaluate the quality of the column.
We prepared a test mix with naphthalene/anthracene (5/0.5 mg/ml hexane) and run it with hexane 100% @10 and 20 ml/min
The result was that the resolution and peak shape was "poor" with injection of 1 or 2 ml. Plate count was about N=900 which would lead to a plate height of H=255 µm (corrs. h=6.4 (for 40µm particles))
Injecting anthracene for its own in a 1:10 dilution gave better peakshapes and about 2000 plates (H=115µm, h=2.88)
Beside of this the k' values don't correspond to those obtained by short TLC evaluation (Merck HPTLC plates, Si60).
Don't have the values by hand right now (think the Rf were about 0.5 and 0.7). On MPLC the first peak came out virtually with t0, followed by anthracene.
We then tried another test mix with phenone homologues (aceto-, propio-, butyro-) and hexane/t-butyl methyl ether 95/5.
Propio- and Butyro couldn't be separated good enough, so we prepared a solution of only aceto- and butyrophenone, with some toluene as t0 marker, dissolved in hexan 100%. b(phenones)= about 12mg/ml total.
Injecting of 1 ml gave a satisfying resolution, the k values were about 0.29 and 0.75, t0 was 4.8min, flow 21 ml/min. N was about 1200 (for the second peak.)
Injection of 0.5 ml the plate count rises to about 1800.
Injection of 2 ml, the plate count droped to about 800.
Beside of the plate counts there are these small k values. Evaluated by TLC we would have excpected k's of 1 and 1.6...
Now I wonder if someone can unravel these mysteries...
What effects do I see?
- Volume-overload, even if sample is disolved in eluent? Already with 1 ml or 2 ml?
- Bandspreading before and after column? (bandspreading volume was evaluated to be about 12 ml (w4.4% method))
Which plate count is "true", should be used for the estimation if a separation can be done with this colume?
How do/can I get these significant deviations in k values for silica (TLC vs MPLC)?
Where should I further "shoot troubles", column or flowpath?
Are these observations "normal" for MPLC?
What are some rule of thumbs for MPLC?
- reduced plate heights?
- injection volumes related to the column volume?
- mass load /g silica
- tests for evaluation, if the column can be used or bette should be repacked?
Thank you for your replies