EPA 524.2 Chlorofluorocarbons
Posted: Fri Dec 14, 2018 5:41 pm
by mduplease32
I run EPA 524.2, purge and trap concentrator connected to GCMS. I've noticed lately that I have been having a problem with the chlorofluorocarbons. Whenever I calibrate, my curves for dichlorodifluoromethane and trichlorofluoromethane come out great (<20% RSD) However, when I run a CCV the concentrations for these analytes come out much higher and I fail the 30% acceptance criteria. Any ideas?
Re: EPA 524.2 Chlorofluorocarbons
Posted: Mon Dec 17, 2018 5:05 pm
by Steve Reimer
Is the ccv made from the same stock/daily as the cal? If gasses come out high the cal standards were low. How was your 2nd source check?
Posted: Tue Dec 18, 2018 2:15 pm
by Bigbear
DOn't forget to check the fluorobenzene, if it's low.....
Posted: Tue Dec 18, 2018 3:42 pm
by James_Ball
I use the MiniInert caps on 5ml reaction vials to store my volatiles standards to prevent loss of the lighter analytes such as these. If your standard is a few days old and you have to open the cap each time you make a dilution you will lose the gasses very quickly, within a couple days. I also use the standard still cold, I don't allow it to warm completely to room temperature. The reasoning behind this is that the small volume of methanol you pull up into a gastight syringe will equilibrate to the temperature of the syringe barrel within a couple seconds, so just pull up the amount you need, wait a few seconds for equilibration and you have an accurate amount. Plus keeping the standard cold reduces loss of analytes over time.
https://www.sigmaaldrich.com/analytical ... alves.html
Posted: Fri Jan 04, 2019 7:37 pm
by mduplease32
The CCV is made from the same stock as the cal. I make my cal standards from brand new stock. The 2nd source check (QCS) comes out high as well.
Posted: Fri Jan 04, 2019 7:39 pm
by mduplease32
But wouldn't you expect to see the concentrations coming out lower if that were the case? The concentrations are increasing over time.