Cleanup for determination of pesticides by GC
Posted: Mon Sep 08, 2008 6:56 pm
by astroguille
Hi people!
I work on a GC/MSD (6890/5973) for determination of pesticides in fruits, vegetables, water and soil.
Which is the best and cheapest cleanup?
Thank you
Re: Cleanup for determination of pesticides by GC
Posted: Mon Sep 08, 2008 7:16 pm
by paulolab
Hi people!
Thank you
Try to use quecher's method. It's the most inexpensive method.
Posted: Tue Sep 09, 2008 1:04 am
by AICMM
astroguille,
I have to admit that quechers looks very interesting. I just looked it up tonight and I am going to read more about it over the coming days.
In the "old" days I would recommend gel permeation as a good way to remove fats and oils from pesticide extracts. Uses a lot of solvent but can certainly be built inexpensively in any lab (about 2-2.5 K without a UV detector and can be made to work with any used HPLC UV 254 detector.) Good thing about this cleanup is it is size based so minimal worries about pesticide/cleaning agent interactions. I can provide more info or you can look it up on the Accustandard or EPA website.
Best regards.
Posted: Wed Sep 10, 2008 2:21 pm
by astroguille
Thanks people!
I entered in the EPA website, but I didn't find anything.
AICMM, can you provide more info, please?
Posted: Wed Sep 10, 2008 3:29 pm
by oscarBAL
Hello Astroguille;
There exist approved metohd for Quechers at AOAC.
And FDA.
Regards.
Oscar
Posted: Wed Sep 10, 2008 7:47 pm
by jh1
Hello Astroguille;
And FDA.
Regards.
Oscar
Also quechers.com has methods and information listed.
Posted: Wed Sep 10, 2008 8:05 pm
by zokitano
Also, I think this is an useful article in LCGC magazine regarding QuECHeRS:
http://www.lcgceurope.com/lcgceurope/Co ... ail/473128
Regards
Posted: Thu Sep 11, 2008 10:30 am
by AICMM
astroguille,
http://www.accustandard.com/asi/epa_downloads.php3
http://www.epa.gov/osw/hazard/testmetho ... series.htm
Two web sites that have Method 3640 which discusses GPC cleanup.
If after reviewing the method you want to know how to build it, please send me an e-mail through my web site and I will send you guidance.