i ame working in clinical diagnostics so robustness of the hardware is very important.
Advertisement
Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.
We are thinking of adding a newer LCMSMS to the lab and since Agilent has said they will soon discontinue the 6460 we are instead looking at the 6470. Can the 6470 still be purchased without the iFunnel system? From reading the specs that is an extra add-on so if we don't add that should it be more comparable to the 6460? Our other instrument is the Sciex API3200 so I know for certain we can not just transfer methods, but am hoping that the much smaller injections will help keep it clean.Hi,
Let me clarify..
The reason newer instruments seem less robust is that they are more sensitive.. Especially funnel based systems compared to older skimmer types..
If you had gone with a 6460 compared to your old waters you would most likely have found it very robust.. But sesitivity comes at a price.
Regarding sample prep issues. Many of these are more related to the LC in front.. With new high end MS people will also usually have high end LC systems with narrower capillaries. This makes them more sensitive to poor sample prep and mobile phases. But the MS rarely stands alone so "bad sensitivity" is usually blamed on the MS but often it comes from the LC.
Hey James, just a small comment from experience: you will notice that all cobalamins (B12 vitamers) ionize much better as [M+H]2+ in ESI, which is in the 600-700 m/z range.In the past we have also done some Vitamin B12 analysis and I believe that is pushing the 1400 mass limit.
Thanks for the tip. We tried the perfluorinated method for the last UCMR3 and just could not get down to the detection limits needed for that with the 3200, but then a 6470 is at least 10x more sensitive so that should work well. Agilent has an application for the 6495 that is a direct injection of the water sample with no extraction/concentration, but if the 6470 would be more robust, I think I would do the extraction first.Hey James, just a small comment from experience: you will notice that all cobalamins (B12 vitamers) ionize much better as [M+H]2+ in ESI, which is in the 600-700 m/z range.In the past we have also done some Vitamin B12 analysis and I believe that is pushing the 1400 mass limit.
About perfluorinated compounds: I don't know which instrumental limits you need to reach, but we were able to validate surface- and waste waters on our older LCMS instrument. Lowest calibration point is 1 µg/L with 10µL injection. MS-wise these compounds are straightforward/easy to develop: they all ionize as [M-H]- in ESI, sensitivity is quite good.