Normalisation or calibration

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi All,

I was wondering about how to justify a normalisation procedure for related substance to the regulatories.

When I inject a sample solution, I use a 1% solution to calculate the%unknown related substance relative to the main peak in the 1% reference.
But if I look at the %area it's equal to the calculated %.
So I was thinking that normalisation can be used, and is even more precise because it's not influenced by weighing errors.
But how do I justify this approach in my report to the regulatories?

Any ideas?

Thanks for your time

Ace

You have to demonstrate that you have the requisite 3+ orders of magnitude linear range (i.e., all the peaks are in the linear range of the method). That begs the question of any assay of this type: the assumption that all peaks have the same molar absorbance.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Hi Tom,

the linearity isn't a problem, I already tested it from 0.01% to 130%..... for my main component. But I also have some small peaks, for which it is accepted to use a 1% reference by the pharmaceutical regulatories.
But I can't purify all of them and determine their molecular absorbance.
Or is there another way?

Ace
3 posts Page 1 of 1

Who is online

In total there are 13 users online :: 2 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Baidu [Spider] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry