7694 HS Transfer line Temp equilibration

Posted: Wed Sep 03, 2008 7:21 pm
by contributor
Does anyone know why the transfer line temperature is fluctuating between 109 and 102° when the setpoint is 105°. It cant reach the setpoint. I baked the line at 180° overnight but that hasn't worked. I checked the sensor resistance like the manual suggests and we have replaced the transfer line with another one that was laying around. Nothing has helped. It will maintain 102° and 110°.

Thanks
Alan

Posted: Thu Sep 04, 2008 4:28 pm
by danicrd
the problem can be in the temp controller if the resistance is ok..or maybe there is non insulated zone that disperd heat
good luck