What's your SIM dwell time, how many peaks are in a typical SIM window and have you corrected your SIM ions for mass defect?I'm running EPA 8270 and, as the title explains, I'm getting fine chromatography on scan mode but crummy peaks on SIM.
any thoughts?
I'd recommend shortening the dwell time to 30 for every ion in each group, see how that looks then go from there.Hi there,
dwell time is 100 and typically I have 8-12 ions per group.
I'm not sure how to do a correction. The manual is very limited when it comes to explaining SIM.
Thanks.