Sizeexclu chr,resolutn..18kd intactprotein and 15 kd degradn

Discussions about gel permeation chromatography / gel filtration chromatography / size exclusion chromatography

4 posts Page 1 of 1
hii..
i need to develop a size exclusion chromatography for resolution of 18 kd intact protein and 15 and 10 kd degraded products of the protein.
any suggestion for columns for the range for which i can expect resolution?
thanks!!
Explore!

Protein-Pak 125 from Waters.

Protein-Pak 125 from Waters.
i have been using this column.. but my 19 kda protein.. and buffer/diluent peak are very near to each other.. no way i can see my 14 and 10 kda protein fragments between them.. any suggestions?? thanks!!
Explore!

Several possibilities. The fastest one is to e-mail me a copy of the C-gram. Alternatively, what is the flow rate, what are the retention times of the 19kDa protein and the buffer peak, what is your mobile phase?
4 posts Page 1 of 1

Who is online

In total there are 7 users online :: 0 registered, 0 hidden and 7 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 7 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry