High pH Reverse Phase Column Life
Posted: Wed Aug 27, 2008 6:10 am
Just wanted to start a topic for peoples experiences with reverse phase columns at high pH
I work with small molecules that contain a basic nitrogen, and are best analysed at high pH for both selectivity and sensitivity. To date I have developed a number of methods for these molecules that operate around pH 10, which in comparison to acidic or neutral methods are far superior.
The problem is column lifetime. The first high pH method I developed used the Gemini column …. not good, peaks which had a resolution of 4 initially dropped to below 2 in under 100 injections (35 minute run time). In an attempt to rectify the problem I substituted a XTerra into the method, this doubled column life (200 injections). Please note that the columns are not failing catastrophically in this time, just losing resolution between critical pairs.
As far as I can tell the difference between the two columns above is the Gemini uses a ‘protective’ polymeric layer to protect the silica, while the Xterra is a crosslinked hybrid particle. I presume that once the protective layer on the Gemini particle is breached, the column dies quite quickly, whereas the hybrid particle is more stable.
I have now tried XBridge columns with the hybrid BEH particle, and have increased column life to 300 injections. This was also helped by a reduction in column temperature (now 30 deg C).
Can anyone share their experiences under similar conditions?, has anyone tried the new Gemini column?, or can recommend a good high pH column. I also have not tried the Agilent Extend columns, although am sceptical about how base resistant these would be.
Perhaps a zirconia or polymeric column would be the way to improve column life, I am curious about the possible loss in efficiency in comparison to silica based columns, and am very interested to hear what peoples experiences with zirconia has been.
In anticipation to requests for more information…. I have tried vendor recommended regeneration procedures, and no, I do not use a guard column. I am fortunate enough to be analysing nice clean pharmaceutical actives. The mobile phase is ammonium formate for one method and triethylamine bicarbonate for another, with no difference in column life between the two.
Thanks
I work with small molecules that contain a basic nitrogen, and are best analysed at high pH for both selectivity and sensitivity. To date I have developed a number of methods for these molecules that operate around pH 10, which in comparison to acidic or neutral methods are far superior.
The problem is column lifetime. The first high pH method I developed used the Gemini column …. not good, peaks which had a resolution of 4 initially dropped to below 2 in under 100 injections (35 minute run time). In an attempt to rectify the problem I substituted a XTerra into the method, this doubled column life (200 injections). Please note that the columns are not failing catastrophically in this time, just losing resolution between critical pairs.
As far as I can tell the difference between the two columns above is the Gemini uses a ‘protective’ polymeric layer to protect the silica, while the Xterra is a crosslinked hybrid particle. I presume that once the protective layer on the Gemini particle is breached, the column dies quite quickly, whereas the hybrid particle is more stable.
I have now tried XBridge columns with the hybrid BEH particle, and have increased column life to 300 injections. This was also helped by a reduction in column temperature (now 30 deg C).
Can anyone share their experiences under similar conditions?, has anyone tried the new Gemini column?, or can recommend a good high pH column. I also have not tried the Agilent Extend columns, although am sceptical about how base resistant these would be.
Perhaps a zirconia or polymeric column would be the way to improve column life, I am curious about the possible loss in efficiency in comparison to silica based columns, and am very interested to hear what peoples experiences with zirconia has been.
In anticipation to requests for more information…. I have tried vendor recommended regeneration procedures, and no, I do not use a guard column. I am fortunate enough to be analysing nice clean pharmaceutical actives. The mobile phase is ammonium formate for one method and triethylamine bicarbonate for another, with no difference in column life between the two.
Thanks