Posted: Mon Aug 18, 2008 7:55 pm
by LWC
Hello,
I had a ISTD contamination problem on my GC/MS, I have replaced the liner, septa, cut a small piece off of the injection side of the column, baked out the column, and done solvent rinses, but nothing seems to work. Does anyone have anysuggestions as to what I should try next. The ISTD is 1-Chloro,octane.
Thanks.
Posted: Tue Aug 19, 2008 6:01 am
by frage
Hello!
If you have a autosample change sirnga and change vials.
And then put injector and column high temperature for a night (check the t max).
And we hope good!
Posted: Tue Aug 19, 2008 7:31 am
by Peter Apps
Some questions first:
does the contamination appear as a peak, or as background ?
If you run only the GC-MS programme, with no injection, do you still see the contamination ?
If you do a dummy injection (i.e. nothing in the syringe) do you still see the contamination ?
If you inject pure solvent do you still see the contamination, and is it still there with different solvents ?
Peter
Re: contamined column Peter Apps
Posted: Tue Aug 19, 2008 12:41 pm
by LWC
Yes the comtamination is a peak.
I am working on the no injection now, if I do an "air" injection yes it is still there.
Yes at solvent injection still shows the peak.
I am having a service tech come out tomorrow to address the issue, so I will let you guys know the out come of the problem for furture reference.
Thanks
Re: contamined column Peter Apps
Posted: Fri Sep 05, 2008 5:26 pm
by paulolab
It was a durty liner?
Glass wool contaminated?
Contaminated column
Posted: Fri Sep 05, 2008 6:34 pm
by LWC
Sorry I never got back to this I have been slammed at work. Anyway the injection port ended up having to be replaced along with column. All the parts of my autosampler had to be cleaned as well.