Varian Pump 9012/9002 or 9010 fault 68

Posted: Tue Oct 16, 2018 6:17 pm
by kyec
Hello, Does anyone have experience in diagnosing a fault 68 Pump Stall on a Varian 9010 or 9012 pump? The operators manual suggests to check hydraulics for clogged line. I am getting mobile phase flow through the column. The motor/piston make three turns and then stops with this fault. I cannot see anything hindering the movement of the motor and piston. I swapped out the optical sensor with a working unit and this did not help. The pressure goes no higher than 20 barr. Thanks!

Re: Varian Pump 9012/9002 or 9010 fault 68

Posted: Tue Oct 30, 2018 11:55 am
by gaumit
Hi,
Do you check plunger seal ? (or try to change new one plunger seal first)
Sometime system fault if system pressure do not meet requirement,

Re: Varian Pump 9012/9002 or 9010 fault 68

Posted: Tue Oct 30, 2018 2:18 pm
by uzman
Make sure that upper pressure limit of the pump adjusted more than 20 bar !