GC-MSD connection issue

Posted: Tue Oct 16, 2018 12:46 pm
by linhong
Hi everyone,

I have some problems with Agilent GC-MSD (5795 MSD). When I run samples, the instrument status became offline after the Prerun programme started, then it stopped after the first sample. I had to restart the software to run the next sample, no matter I run samples in a sequence or one by one. I tried to create a new sequence, use new methods, and restart the computer, but it still has this problem. Did anyone have similar problems before?

Re: GC-MSD connection issue

Posted: Tue Oct 16, 2018 2:43 pm
by Steve Reimer
Version D02.xx? E?
Have you checked or redone the instrument configuration?

Re: GC-MSD connection issue

Posted: Tue Oct 16, 2018 8:04 pm
by Trishia
Does it say which part of the system went offline, or it's everything (GC, MSD, potential autosampler)? Anything relevant in the current logbook?

Re: GC-MSD connection issue

Posted: Wed Oct 17, 2018 2:17 pm
by gcguy
I have seen similar problems, I think the network cards get locked up. I have found the best way to fix this is to power cycle everything including the instruments. When doing this, physically pull the power plugs out for 20 seconds. This resets the network cards. Most network cards can stay powered up even when the PC is switched off.

GCguy