Agilent APG/ERI remote - poor triggering precision

Posted: Tue Oct 09, 2018 9:17 am
by MaxSpecs
Hi,

Would anyone have experience with Agilent APG or ERI remote triggering?
I'm triggering a 1290 kit via ERI (an external injection valve sending output - completed turn - signal to trigger the LC) and seeing some unwanted variation in retention times as measured on t0 ranging up to 0.8 s. This may not sound like much but I work with ultra fast methods and that variability is effectively ruining my runs. Interestingly, when the same method is executed on the autosampler the injection is very precise, low RSD on retention times.
I'm suspecting that this is due to some internal software delays, i.e. variable delay between the moment the trigger is registered on the terminal and when the run is actually initiated by the relevant software routine and t0 is marked.

Are there any recommended ways of triggering the Agilent kit with much greater precision similar to the autosampler operations?

Many thanks

Re: Agilent APG/ERI remote - poor triggering precision

Posted: Mon Oct 22, 2018 4:49 pm
by leadazide
One possible reason can be that when you trigger with external signal the pump and AS don't sync the piston stroke.. That could explain why it works when using the Agilent AS..

What valve are you using to trigger the start?

What software are you using?