I have a very interesting problem. I am developing an assay/impurity method w/ following parameters.
column - chirex 3125 d-penicillamine 250x4.6 5u
MP - 1mM Cu(II) Acetate (aq)
flow - 1 ml/min
Col temp - 25C thermostatted
inj vol - 20 uL of 2 mg/mL sample
Detection - 230 nm, VWD
Run time - 40 minutes
Sample: N-substituted L-arginine
HPLC - Agilent 1100
Injection of the 2 mg/mL sample leads to a baseline disruption 7 hrs later.
I have confirmed this w/ duplicate experimental studies where five samples were injected using above conditions, then the LC system switched to a 480 minute run time method. In each case, the baseline drops ~ -10 mAU 7 hours laters, then recovers over 40 minutes corresponding to the assay method run time. The initial drop is not sharp, but steep enough to make integration of trace peaks difficult.
Comments
Thanks,
Robotjock