Can anyone tell me why do we get a void volume peak or a peak with negative absorbance at the start of each run in hplc?
Usually this is because the matrix you introduce the sample in has a lower response than the eluent being run. This slug of liquid does not interact enough with the mobile phase to concentrate to the same consistancy.
It comes from pressure impuls when you do the injection and you can see what ever compound is not retarded to the packing material. Pressure impuls when doing gradient elution can cause a slight change in the composition of your MP in the detector cell and can change the RI for a short moment. thant can give you a negative "Peak".
Gerhard Kratz, Kratz_Gerhard@web.de
Also, difference in the refractive index between air and/or the sample solvent and the mobile phase you're using may contribute to a "void peak" that points in the opposite direction of your peaks of interest.
MattM
Thanks
5 posts Page 1 of 1

Who is online

In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry