Posted: Thu Aug 07, 2008 2:13 pm
So, I have a Waters Alliance system sitting in my lab that my group wants to dedicate solely to our projects, rather than split time with people in the other labs. Problem is, it has been dormant for about 2 years. So here's the story:
I have the LC moved from its current location to my lab, and begin to hook everything up. Computer still works, Empower is still loaded and I even manage to track down the username and password. (That was a project itself). I hook up the HPLC and everything comes online. No malfunctions which makes me happy.
I fill solvent bottles with ACN and Water so I can get liquid running through the system again. I dry prime by wetting the lines with a syringe and switching to 100% composition.
Long story short, no matter how long I dry prime, in line A i get only about 145 psi, and 45 psi in line B. The pressure doesn't remain very constant either, it will jump around signficantly.
I contacted waters, and they're supposed to have a service engineer call me (The guy I originally tried to call no longer works at Waters) so they can set me up with a service plan and come out and do a PM.
In the mean time though, is there anything I am missing? Do I just need to continue to prime? In the past when i had low pressure on a system It was due to a leak somewhere, or my solvent bottle ran dry overnight and there was air in the system. That however was on an Agilent and those leaks were easy to trace. Is it possible there is a leak I can't see? Is there anything else I can try?
Thanks again for the help and support.
Tim
I have the LC moved from its current location to my lab, and begin to hook everything up. Computer still works, Empower is still loaded and I even manage to track down the username and password. (That was a project itself). I hook up the HPLC and everything comes online. No malfunctions which makes me happy.
I fill solvent bottles with ACN and Water so I can get liquid running through the system again. I dry prime by wetting the lines with a syringe and switching to 100% composition.
Long story short, no matter how long I dry prime, in line A i get only about 145 psi, and 45 psi in line B. The pressure doesn't remain very constant either, it will jump around signficantly.
I contacted waters, and they're supposed to have a service engineer call me (The guy I originally tried to call no longer works at Waters) so they can set me up with a service plan and come out and do a PM.
In the mean time though, is there anything I am missing? Do I just need to continue to prime? In the past when i had low pressure on a system It was due to a leak somewhere, or my solvent bottle ran dry overnight and there was air in the system. That however was on an Agilent and those leaks were easy to trace. Is it possible there is a leak I can't see? Is there anything else I can try?
Thanks again for the help and support.
Tim