I am running a 6890 Agilent GC with splitless flow. I recently dropped my run time to 14.25 minutes for pesticides and the chromatography looked great. However, suddenly, I began getting what appear to be double peaks for the same analyte: a large normal peak and a small hump attached following the initial peak. I can only assume this second small shoulder is part of the initial "parent" analyte, since it is occurring for each peak in the first half of the run. I switched back to my normal run of 24 minutes to see if there were any change--there was not. Has anyone experienced this, or are there any suggestions in correcting the problem. It is making integration and calibration impossible--I cannot rely on what may be part of the same analyte. Thanks in advance for any advice!
-
- Posts: 176
- Joined: Mon Mar 14, 2005 7:01 pm
What is your old and new oven temperature pgrm?
What is your solvent and column?
What is your solvent and column?
-
- Posts: 64
- Joined: Tue Jan 15, 2008 4:22 pm
Which pesticide and have you purchased new standards recently?
This secondary peak could simply be a different isomer of the pesticide.
This secondary peak could simply be a different isomer of the pesticide.
-
- Posts: 50
- Joined: Fri Aug 01, 2008 9:41 am
Hi
I'm sorry about my (very) bad english
If appears doublepic in many pics of your cromatogram it's possible a problem with the injection.
I'm sorry about my (very) bad english
If appears doublepic in many pics of your cromatogram it's possible a problem with the injection.
Francesc
Sorry for my english
Sorry for my english
-
- Posts: 19
- Joined: Tue Jul 15, 2008 4:30 pm
probably the column was dead or overloading. Check with new column and smaller injection volume.
-
- Posts: 101
- Joined: Wed Sep 12, 2007 12:20 pm
try lower initial temp of column: the problem is simply injection problem
good luck
good luck
------------------
daniele
------------------
daniele
------------------
-
- Posts: 50
- Joined: Fri Jun 17, 2005 7:26 pm
I have seen this with a Thermo Focus GC and AS 3000 autosampler. Simply increasing the injection needle dwell time to 2 seconds solved my problem. Never noticed anything like that with the Agilent autosampler which seems to have the fastest injection i know of.
-
- Posts: 21
- Joined: Tue Aug 05, 2008 11:17 am
Hello!
test see if the gold Plated is clean?
P/N 18740-20885
test see if the gold Plated is clean?
P/N 18740-20885
8 posts
Page 1 of 1
Who is online
In total there are 174 users online :: 1 registered, 0 hidden and 173 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google [Bot] and 173 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Google [Bot] and 173 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science