Double Peaks of the Same Analyte

Discussions about GC and other "gas phase" separation techniques.

8 posts Page 1 of 1
I am running a 6890 Agilent GC with splitless flow. I recently dropped my run time to 14.25 minutes for pesticides and the chromatography looked great. However, suddenly, I began getting what appear to be double peaks for the same analyte: a large normal peak and a small hump attached following the initial peak. I can only assume this second small shoulder is part of the initial "parent" analyte, since it is occurring for each peak in the first half of the run. I switched back to my normal run of 24 minutes to see if there were any change--there was not. Has anyone experienced this, or are there any suggestions in correcting the problem. It is making integration and calibration impossible--I cannot rely on what may be part of the same analyte. Thanks in advance for any advice!

What is your old and new oven temperature pgrm?
What is your solvent and column?

Which pesticide and have you purchased new standards recently?

This secondary peak could simply be a different isomer of the pesticide.

Hi
I'm sorry about my (very) bad english :(
If appears doublepic in many pics of your cromatogram it's possible a problem with the injection.
Francesc

Sorry for my english

probably the column was dead or overloading. Check with new column and smaller injection volume.

try lower initial temp of column: the problem is simply injection problem
good luck
------------------
daniele
------------------

I have seen this with a Thermo Focus GC and AS 3000 autosampler. Simply increasing the injection needle dwell time to 2 seconds solved my problem. Never noticed anything like that with the Agilent autosampler which seems to have the fastest injection i know of.

Hello!
test see if the gold Plated is clean?

P/N 18740-20885
8 posts Page 1 of 1

Who is online

In total there are 174 users online :: 1 registered, 0 hidden and 173 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot] and 173 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry