krickos » Fri Jun 27, 2008 7:17 am
Willy,
What are the headaches if linearity is evaluated from QL to 120% specification limit?
I have done residual solvent linearity this way for some time, I got a good linear regression line and 0.99+ correlation coefficient. It seems fine for me.
Terry
What I think Willy is trying to say and what I also agree with in general is that, sure the linearity many times is accepteble down to the DL, but from a Range perspective the accuracy and precision usually detoriates quite a bit from the QL from time to time. So practically there is not always a need to push the linearity invastigation to far down in concentrations.
I have one/two other comment regarding how far down in concentration you want to push your validation.
For organic impurities it is somewhat easier as ICH Q3a+b quite cleary tells what the expect, but for residual solvents (ICH Q3C) it may become more complicated especially on the global market. Class 3 solvents rarely complicates things, but class 2 solvents tend to. Especially in EU and Japan regulators tend to sharpen the limit below the ICH Q3C limit based on the batch data submited. Just had a discussion with another company where EU dropped the toluene limit from 890ppm to 100ppm for one of their drug substances.
Also if your submiting or want to change a specification limit for a class 2/1 solvent in Europe, EU (EMEA) has published an annex to the ICHQ3C guideline with clearification on expections. For example what data/discussions is needed to rule out a class 2 solvent from the specification.
This can also have impact on how far you want to push the validation. Unfourtunatly my PC is currently not working as expected so I can not provide a link at this time but you can find it on
http://emea.europa.eu/, continue to human medicines and look in quality.
The annex came into operation in January 2005 and is labled:
CPMP/QWP/450/03
EMEA/CVMP/511/03
in the top right corner.
Cheers Chris