Posted: Tue Jul 29, 2008 12:20 pm
Hello at all
My system is Varian 3900/saturn2000 coupled with Tekmar lsc3100 purge e trap, for 8260 analysis.
Recently we have much problem with IS peak: when we change the trap the peaks are sharp, but only after 100 injection they has big tailing. The trap is vocarb 3000, desorbed at 250°c and baked out at 260°C
Direct injection was good, so I think gc/ms area works fine.
The problem seems trap related, but we can't change a trap every batch!
Has anyone experienced similar problem?
Thank you