Longer description:
Agilent 1260, DAD G4212
Pretty straighforward program: HPLC water with pH 2.5, no salts - Methanol gradient:
99:1 (hold for 2 min) change to 20:80 over 9 min, return to initial composition.
Regular baseline has a slight elevation (or 'bump') in the middle.
During the analisys gradual shift of baseline occurs toward slight downward slope in the middle. Then during one of the runs a sudden 'jump' of the baseline, and on the next chromatogram, where initially was a little 'bump' due to gradient, is now a huge drop (better see the pictures). The baseline profile stays about the same after this. Turns out the problem is with the optic cell. Changing the cell to another solves the problem, and strange downward profile reproduced with the problematic cell on another HPLC.
The problem is not just weird baseline form, it is also more noisy and affects minor analyzed components.
We tried to purge the problematic cell with HPLC water heated at 60 C, both forward and backward flow - it didn't help. Any ideas as to what could be the problem and what can be done about it?
Thank you.