Good evening, Tubish81,
I see. Accuracy is dependent upon the expected goal for the method, often for a potency method, it may be 98.0 - 102.0 percent for a major component, for example. Perhaps this will help:
https://www.fda.gov/downloads/drugs/gui ... 134409.pdf
and/or this, a discussion in this Forum a while ago, HPLCAddict made the following salient points:
I [HPLCAddict] don't think that there are any official guidelines out there which give definite acceptance criteria for accuracy. And this is actually good, because acceptance criteria for accuracy (and any other validation parameter) can be wildly different depending on the purpose of the respective analytical method. Regarding the +-2% you mentioned, this is quite a common criterion for validation of assay tests in the pharmaceutical field. But for related substances, things are VERY different! Just think of it: at the 0.1% level defining an acceptable range of +-2% would mean you'd have to hit the 0.098-0.102% range with your recovery - impossible! I'd say there's no pharmaceutical related substance method out there which is precise and accurate enough to distinguish between these concentrations. An actually it doesn't have to.
I've seen a lot of validations of related substances methods, and usual acceptance criteria for accuracy via recovery are 90-110% or even 80-120%, especially if you look at lower concentrations close to the LOQ.
I agree with HPLCAddict wholeheartedly. Perhaps also this will help:
http://www.eoma.aoac.org/app_f.pdf
As to Robustness, this would have to be determined during method development on a case-to-case basis, that is, there are no pre-set criteria (perhaps your firm suggests some in an SOP?). As for criteria to examine in robustness, there is a list in this text you may follow (in Chapter 2):
Validation Of Analytical Methods For Pharmaceutical Analysis by Oona McPolin,
ISBN-13: 9780956152817, Pub. Date: 05/05/2009, Publisher: Mourne Training Services.