HPLC-tocopherols-standards
Posted: Fri Jul 25, 2008 1:05 pm
by dolenjc2000
Hi guys!
Will be running analyses of tocos and tocotrienols. I purchased all four standards alpha, beta (in hexane solution), gamma, delta toco at Sigma (cheap) in order to dilute them with methanol and then measure absorbtion. With this i will be able to say the finnal concentration of specific toco. How should I prepare my standard, what is the equation for concentration determination, how should i proceed with beta toco (in hexane solution). I am determinating tocos in vegetable oil.
Thnx
Robert
Posted: Sun Jul 27, 2008 2:45 am
by tom jupille
There is a section on calibration/dilution calculations in our tutorial "Getting Started in HPLC". Look here:
http://www.lcresources.com/resources/ge ... ction4.htm
AOCS method
Posted: Mon Aug 04, 2008 12:23 am
by ljc
Check with the AOCS for their standard method (sounds like you may have part of it, since the approach you're taking to calibrate sounds like it was taken from it). You make a known wt/Vol solution with each standard in MeOH, measure their absorbances, calculate the actual concentration using published molar absorbtivities, then evaporate the MeOH to dryness and take up the residue in hexane for subsequent NP HPLC analysis. Pretty tedious IMHO so we instead use NMR to determine purity of a secondary standard and just weigh them out, using the purity value to correct for the actual concentration.
more info
Posted: Mon Aug 04, 2008 12:36 am
by ljc
Dolenjc2000, I don't know how to steer you to a post I made a few yrs ago so I just copied it here...
The HPLC method for Vit E including tocos and T3s is well documented, check the AOCS. There are minor variations in the literature, but it is a normal phase method using either a DIOL or silica column and FLD at 290 nm excitation/330 nm emission. We do this analysis routinely using the following conditions:
Supelco DIOL 250x4.6 mm 5 micron at 35 deg C (Supelco calls it a "diol-like" column, and I've found it to be very robust under these conditions, yet different than other DIOL columns that are out there)
Eluent: 0.5% isopropyl alcohol in heptane, isocratic at 1 mL/min
Inj volume: 20 uL (you can adjust of course, but you have to be careful not to go too big unless you're using straight heptane as the sample solvent).
Detector: FLD excitation: 295 nm emission 330 nm
The vitamers elute in less than 20 mins in the order alpha, alphaT3, gamma, gammaT3, delta, delta T3. The betas elute between the alphas and gammas I think, but are very low in concentration, at least in rice-derived oils. The glyderides elute before the alpha toco and don't interfere anyway using FLD (though it's nice to have pure peaks anyway)
Calibration: Here's the catch: the "standards" for tocos or T3s are hard to get (Merck, Calbiotech, etc) and are extremely expensive, and are in such small ampules (50 mg or so) that it's hard to make solutions unless you use the AOCS method. This calls for making a solution with a candidate standard in methanol and measuring its UV absorbance, then back-calculating the "true" concentration from published molar absorbtivities (which were obtained in MeOH). You then add an aliquot of the meoh solution, evap to dryness, and take the residue up in your normal phase solvent. Theoretically you know the mass in the MeOH aliquot from the calculated concentration, the final volume of the heptane you dissolved the residue in, and therefore its concentraton. This is tedious especially since std solutions have a limited lifetime. You can do this with inexpensive alpha and delta from Aldrich, just run a separation on them to be sure they're at least pure in the vitamer of interest. Finally, the accepted practice is to set the response factors of each T3 to be the same as their corresponding tocos. This makes sense since the chroman ring is the part that either absorbs or fluoresces.
Your sample prep for edible oils is trivial: just dissolve in heptane
The remaining details about dilution factor, how many levels to calibrate, etc have to be determined by you based on the types of samples you have.
There are also reports of reverse phase methods but with oils, the normal phase seems better since you have the perfect sample diluent already as the mobile phase, and the vitamers are separated isocratically very quickly.