Posted: Thu Jul 24, 2008 2:35 pm
by leticiahf
Hi
I am facing a problem in CG (FID) . When I am going to execute a method, the air flow, which would be 400 mL/min, does not reach this value, remaining in 0, and the message " Det APC1 air leaks" appears. What is happening? How can I solve it?

Sincerely
Letícia

Posted: Thu Jul 24, 2008 3:47 pm
by Ron
The error message simple means the electronic flow control (APC) can't reach the proper pressure or flow reading. This is not always due to a leak, it sounds like you may not have flow into the detector.

Posted: Thu Jul 24, 2008 4:52 pm
by leticiahf
so, how to solve the flow into the detector?

Posted: Thu Jul 24, 2008 5:13 pm
by chromatographer1
Your air supply may be turned off or a tank may be empty.

Check.

best wishes,