GC/MS is flooded with mass 11.6 ???
Posted: Mon Aug 27, 2018 1:51 am
by thohry
Anyone ever have the problem of GC/MS having large peak at 11.6 amu ?
I still can see peaks 69, 219 and 502, but the peakwidth is more than 10 times and normal.
Thanks for any suggestion.
Re: GC/MS is flooded with mass 11.6 ???
Posted: Mon Aug 27, 2018 7:51 pm
by JMB
Are you scanning down to m/z 10?
You will burn out the multiplier with the m/ 18, 28 and 32 signals. People generally set scan range at m/z 40-540 or similar in EI mode.
Are the m/z 69, 219 and 512 ions ALL at 10x normal peak width?
Tuning lenses may need cleaning, and/ or resolution potentiometers may need resetting.
JMB
Posted: Tue Aug 28, 2018 1:44 am
by thohry
JMB
Thanks for your input.
The peaks 69, 219 and 502 have quite normal shape, only too large !.
We cleaned the source already.
Posted: Wed Sep 12, 2018 2:12 pm
by James_Ball
JMB
Have you ran a peak width calibration in the tune menu? That should fix the wide peak widths, if not then either there is a problem in the quad electronics or you have a really bad leak. The low mass peak you are seeing could be 18m/z from water.
Another thought, make sure you did not get a tank of Nitrogen instead of Helium installed, they have the same fitting so they can be installed by mistake, that would give a massive 28 and possibly 14 mass that could be the 12 that you are seeing.