Empower Summary Report
Posted: Mon Aug 20, 2018 1:48 pm
by ydna1977
Hi ,
Any way of seeing the summary report (% RSD of n=6 for example for precision) before the sample set has been processed
ie all injections complete and now you want to check our sst and standard verification
All this is done using labels and summary report but you have to process sample set first creating a result set
Can you not review peak RSD before processing ?
Re: Empower Summary Report
Posted: Sat Aug 25, 2018 5:55 pm
by vofelyp
Hi
There is no any peak data (Rt, area, height) before processing. If you want checking the SST you should load the injections in the Review and process manually. The peak RSD% can be calculated in Excell or any other stat. sw.
Or the Empower can be programmed to process and check the SST samples and stop the sample set before injecting the real samples when the SST fails. You can find the technical note about it on waters.com