A problem...
I was runing a gc analyses on Agilent 6890 normal untill i changed empty bottle with full bottle H2 (5.0). After that the baseline from 5-6 pA rised do 30-35 pA. I called the gas service and told them my problem. They said that might be the problem beacuse they changed the gas filler. Today they brought me a new bottle of H2 (5.0) and the baseline is the same 35 pA.
Bellow you will see a chromatogram when i was changing the amunt of H2 in the gc. I started from my normal flow 50 ml/min (35 pA) and came to 10 ml/min (the minimum that is able that FID still runs) where baseline was normal for me 5 pA. Is normal that lowing of H2 causes such baseline fall or is it a problem that the H2 of this supplier still is not so pure.
Lowering of other gases (He, air) does not cause such an baselinefall.