Hi,

I ran across a GC-MS report as seen below:
https://ibb.co/ftBd6U

The tester concluded the essential oil was adulterated due to the low TIC count.
How could one estimate the level of dilution just from the resulting report?
I read up and understand that a quantitative analysis could have been done, with the use of an internal standard in a known concentration, or better perhaps with a curve of an IS in various concentrations.
How does one assess non volatile content in a GC-MS?


thanks for the help
My guess is they're assuming it was cut with something non-volatile because they don't see the adulterant on the scan (except for the phthalates). TIC simply refers to the peak area when you're not running in SIM-mode or looking at a specific ion. So what they're saying is that if this were unadulterated the peaks associated with Agarwood essential oil would be bigger. If they're doing headspace analysis, non-volatile additions like sugars or other organics can reduce volatility of organic compounds, resulting in lower concentrations in the headspace and since they're non-volatile they won't show up in the chromatogram themselves.

However I don't know that this is the best way to detect adulterants. Without the QC such as an internal standard like you mentioned and/or a calibration curve, it's hard to say definitively if the low TIC is due to the sample or due to the instrument/method. With this report alone, it would be difficult to estimate the amount of dilution/adulteration with any sort of rigor. I have seen papers where people have used ion suppression as a sign of adulteration (using cachaca to dilute scotch, for example) and they basically made a calibration curve using different levels of adulteration (from pure scotch all the way to pure rum).

Hope that helps
Thank you very much for the reply, I did find it useful.

>>If they're doing headspace analysis, non-volatile additions like sugars or other organics can reduce volatility of organic compounds, resulting in lower concentrations in the headspace and since they're non-volatile they won't show up in the chromatogram themselves.

Sorry I took my time to reply,
I was pretty sure headspace analysis was not used, but I wanted to see if I could ascertain the actual method.
The essential oil was injected with a solvent, methylene chloride.


>>using cachaca to dilute scotch

Hmm, such adulteration didn't even cross my mind....
pfff, is anything pure anymore...?
something like this?
https://www.researchgate.net/publicatio ... ectrometry

thanks
3 posts Page 1 of 1

Who is online

In total there are 8 users online :: 0 registered, 0 hidden and 8 guests (based on users active over the past 5 minutes)
Most users ever online was 599 on Tue Sep 18, 2018 9:27 am

Users browsing this forum: No registered users and 8 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry