Weird dips in diode array baseline

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I'm having a problem with my Diode Array detector (Agilent G1315A). I'm running 52:48 DI H2O:Methanol isocratic, using 210 and 254 nm, both referenced at 350 with a bandwidth of 100. I'm using an Agilent 1100 setup, (binary pump, online degasser, thermostatted column compartment, autosampler). The detector has been doing this for the last week or so, and has done it with two different columns. Any ideas?

Image[/img]

Random dips, or do they always occur at or near the same time in a chromatogram? And do they also show up in blanks?

If they occur at predictable times, the try disabling the reference wavelength. If something in your sample (or blank) absorbs at 350, it will force a negative dip.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

It's somewhat random- it occurs at roughly the same time in the run, but I've also seen it earlier or later. I'll try disabling the reference wavelength and see if that fixes it. Thanks for the tip.
3 posts Page 1 of 1

Who is online

In total there are 22 users online :: 3 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot], Semrush [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry