HCD cell Fragmentation

Posted: Fri Aug 03, 2018 8:55 am
by Valeria
Dear all,
I am working on a method that look for small molecule. Range of m/z from 100 to 600. I am working ddMS2 in confirmation mode but i am not able to obtain fragmentation values for each of the analyte concentration i am working with. The problem is that the parent compound is displayed in the FullMS spectrum but the ion that is detected deos not fragment. I have been playing with the Minimun intensity threshold value and the Maximum Injection time one as well, with no positive outcome. Anyone have some suggestions? Thanks

Re: HCD cell Fragmentation

Posted: Mon Aug 06, 2018 8:15 pm
by leadazide
With very little information you provide I will take a few leaps of assumptions..

You are on an Orbitrap of some sort, most likely a QExactive.

If you get good MS spectra but poor or no msms spectra most likely your quad is dirty, not calibrated or maybe there is a hardware issue.